We have an error when unit is opening, error is " test 4 failed". What does it means and what can we do? Besides we want to learn all test steps.
Test 1
Test 2
Test 3
Test 4
Test 5
Test 6
We have all service manuals but we can not found any information about this steps!
this will change the ma for the kv's and rewrite the data on the a/d converter.this should get rid of the error.otherwise your'e cpu could be corrupted.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Fysares Delali
Ge Amx4 Test 4 failed
July 02, 2012 09:09
We have an error when unit is opening, error is " test 4 failed". What does it means and what can we do? Besides we want to learn all test steps.
Test 1
Test 2
Test 3
Test 4
Test 5
Test 6
We have all service manuals but we can not found any information about this steps!
Thanks...
to rate and post a comment
doug westlake
re: Ge Amx4 Test 4 failed
July 03, 2012 09:46
you probably need to do a full calibration on this system.
to rate and post a comment
Fysares Delali
re: re: Ge Amx4 Test 4 failed
July 04, 2012 09:45
I did calibrate voltmeter, charger and drive calibrating...
to rate and post a comment
doug westlake
re: re: re: Ge Amx4 Test 4 failed
July 05, 2012 12:47
a COMPLETE cal from start to finnish this will rewrite the cal info.
to rate and post a comment
Fysares Delali
re: re: re: re: Ge Amx4 Test 4 failed
July 07, 2012 10:58
Ok I will try it! But what does it means this test!
to rate and post a comment
doug westlake
re: re: re: re: re: Ge Amx4 Test 4 failed
July 08, 2012 11:25
this will change the ma for the kv's and rewrite the data on the a/d converter.this should get rid of the error.otherwise your'e cpu could be corrupted.
to rate and post a comment
Fysares Delali
re: re: re: re: re: re: Ge Amx4 Test 4 failed
July 14, 2012 11:01
I dont understand clearly my friends...
to rate and post a comment
doug westlake
re: re: re: re: re: re: re: Ge Amx4 Test 4 failed
July 15, 2012 03:02
where are you located? i could come out and help.