GE AMX-4 Test 04 - Failed

(4)

Israel Garcia

GE AMX-4 Test 04 - Failed

August 28, 2007 10:48

Hi:
my name is ISrael GarcIa I´m from Mexico,
I write to you since it wanted to know if you could help me with my equipment? (GE AMX 4).

It appears a message when initiating (TEST 04 - FAILED)


I appreciate your attention


ISrael GArcia

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE AMX-4 Test 04 - Failed

August 28, 2007 11:32

Final check-sum error, can appear after a calibration if the unit was not exited from calibration mode correctly. Otherwise something didn't match in the final check-sum for calibration, enter the cal mode for additional error listings.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE AMX-4 Test 04 - Failed

August 28, 2007 03:10

Calibration check-sum error. Can be caused by exiting the calibration phase incorrectly, and the cal data not being saved. Enter the cal. mode for additional / more specific error codes.

Log inor Register

to rate and post a comment