Dear all,
I have faced a problem with table,which in able to move in up/in position(elevation is deactivated).
how can i solve this one.

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: GE brightspeed CT 16 slice

August 10, 2017 11:10

Do you mean you cant move table up/down and in/out?

1 person liked this
Log inor Register

to rate and post a comment

Gufaw Hagos

re: re: GE brightspeed CT 16 slice

August 13, 2017 01:47

yes,to up/down

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: re: re: GE brightspeed CT 16 slice

August 14, 2017 10:25

Ok. Try to make table sinhronization. You need adjustment manual. Test 9 and 10. Give me your email i will send you this manual

1 person liked this
Log inor Register

to rate and post a comment

Gufaw Hagos

re: re: re: re: GE brightspeed CT 16 slice

August 15, 2017 10:33

may mail is gufaw.hagos@gmail.com

Log inor Register

to rate and post a comment

Artyom Kizhakin

re: re: re: re: GE brightspeed CT 16 slice

December 26, 2017 08:42

Hi! can you send manual to dcrub@yandex.ru, we have the same brightspeed and sometimeis it's won't move the table

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: GE brightspeed CT 16 slice

August 13, 2017 01:48

Check your MTCB and servo drives.
Most probably you have a faulty MTCB. If MTCB is ok, first check fuse F01, Then check the two servo drives.

craig@telemed-technologies.com

Log inor Register

to rate and post a comment

(30)
(4)

Martin Martinez

re: GE brightspeed CT 16 slice

August 13, 2017 02:07

Gufaw, Hello.
Please share me the error log to try to help you.
Regards.

Log inor Register

to rate and post a comment

Gufaw Hagos

re: re: GE brightspeed CT 16 slice

August 15, 2017 10:33

here the error log is 265304

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Cbrx2sales Surplus Auctions June 17th

FIRE SALE PRICES- $20, $50, $100 RESERVES- Cbrx2sales closeout auctions take place on DOTmed.com. Bid today on 127 lots of medical equipment. Onsite or virtual inspection available.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (533 lots)
Cbrx2sales
TIMED
Jun 17, 2026 (127 lots)