GE BRX EXPOSURE ERROR

CRAIG WAMBURA

GE BRX EXPOSURE ERROR

January 06, 2012 09:07

I am repairing a GE BRX x-ray system that displays error E12 (mA out of range.) It has similar generator as PROTEUS. All the jumper and dipswitch settings are ok both on the console and HT control board. There is less than 10V at the capacitors. The system preps for exposure (rotation) sometimes and then stops. There is no radiation.
Can anyone help with the situation?

Eng. Craig

Log inor Register

to rate and post a comment

(46)
(3)

Carlos Pons

Re: GE BRX EXPOSURE ERROR

January 07, 2012 10:43

Hello :
You must check, first, in small focus and in large focus, all Ma station, if doesn´t xray for all, you have a problem in Filament driver (you can remove collimator and check if filaments light), if some station works, you must recalibrate Ma.
What do you mean "sometimes" rotation?

Log inor Register

to rate and post a comment

(37)
(4)

carlos mendez

Re: GE BRX EXPOSURE ERROR

January 11, 2012 06:28

some time it dysplay this error when the tube dont shoot xray, could be the KA board

Log inor Register

to rate and post a comment