Hello, we have a problem with a GE CT FXi system,
When is powering up, the system burns component FL7 that is directly to the PDU connector on the Computer board REAR CN1 ASSY (2140183-2), we did replace the board then the system worked fine for two weeks.
The system did 2-3 scan per day, but then the board burn again, same situation. This time we really need to find the root problem before replacing the board again.
Your help is much apriciated.
Enrique Gonzalez
GEG
GE CT FXi problem
October 16, 2013 04:41
Hello, we have a problem with a GE CT FXi system,
When is powering up, the system burns component FL7 that is directly to the PDU connector on the Computer board REAR CN1 ASSY (2140183-2), we did replace the board then the system worked fine for two weeks.
The system did 2-3 scan per day, but then the board burn again, same situation. This time we really need to find the root problem before replacing the board again.
Your help is much apriciated.
Enrique Gonzalez
to rate and post a comment
Carlos Duran
re: GE CT FXi problem
October 16, 2013 11:09
Hello Enrique,
Try to check the ground connections between modules. I think you have a current return.
Let me know if need mote help.
to rate and post a comment
GEG
re: re: GE CT FXi problem
October 17, 2013 08:45
Hi Carlos, really apriciated you input, I will do that, makes sence.
Carlos Hablas español?
Gracias
Enrique González
to rate and post a comment
Carlos Duran
re: re: re: GE CT FXi problem
October 18, 2013 10:12
Yes sir!
If you want, write me to
Ingenieriaclinica@yahoo.es
to rate and post a comment