yuvaraj sankaran

GE CT HILIGHT DETECTOR PROBLEM

February 22, 2007 04:09

Sir
I have found hilight detector MODULES inside the detector OF a HISPEED LXI TO BE DEFECTIVE .could it be attemted at component level repairs.

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: GE CT HILIGHT DETECTOR PROBLEM

February 26, 2007 09:34

This is something I would not attempt to do in the field and especially without a clean room and proper techniques. Unfortunately even though the detector appears to be repairable I would leave that to the experts