I have a GE CT lightspeed and i installed the direct 3d option but there is no application, the icon doesn't appear, already tried load from cold, but still the same, what could be the problem?
The Direct-3D is a prospective recon, only available during acquisition.
You must activate that in scan prescription.
When you finish scan and leave the patient, is not possible redo the 3D, unless you rescan again.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Raymundo Yamamoto
GE CT Lightspeed, need help, software 3d
August 12, 2013 06:43
I have a GE CT lightspeed and i installed the direct 3d option but there is no application, the icon doesn't appear, already tried load from cold, but still the same, what could be the problem?
to rate and post a comment
Carlos Duran
re: GE CT Lightspeed, need help, software 3d
October 18, 2018 01:25
The Direct-3D is a prospective recon, only available during acquisition.
You must activate that in scan prescription.
When you finish scan and leave the patient, is not possible redo the 3D, unless you rescan again.
to rate and post a comment