GE CT Lightspeed plus ( scanning hardware reset failed)

(49)
(20)

Awodem Alain

GE CT Lightspeed plus ( scanning hardware reset failed)

March 09, 2012 06:08

Dear all,
the system failed to rest the hardware and bring out the next error 184128,244518,178574 with The following controllers failed to report initialization status in 120 seconds: OBC.on error 178574.
ping STC and ETC are ok.
ping OBC not respond.
when the scanning hardware reset failed the gantry
display go blank.
any suggestion i will apreciate.

4 people liked this
Log inor Register

to rate and post a comment

Steve Service

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 10, 2012 09:33

Swap the obc and stc. Before installing the boards make sure you change the dip switch settings on each boards. The switches are very small and hard to see. If the switches are not set correctly the system will not recognize the boards.

Log inor Register

to rate and post a comment

(49)
(20)

Awodem Alain

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 10, 2012 05:40

Dear Steve,
when application is down all node OBC, STC and ETC respond to ping; when application go up and scanning hardware reset failed and Gantry diplay go blank with 3 green dots, OBC not respond to ping and stc and etc are ok to ping.
I swap OBC's CPU and LSCOM with STC without any change. I also set the DIP switch for normal config when I swap.

Log inor Register

to rate and post a comment

Steve Service

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 11, 2012 10:32

After you swapped the obc and stc, did you run the flash download tool to flash the controllers? If not, run the flash download update and monitor the download window.

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: Re: GE CT Lightspeed plus ( scanning hardware reset failed)

January 21, 2015 09:48

I also recommend that you check the slip rings for dust. I have had to correct this problem after cleaning the slip rings.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 10, 2012 09:35

Dear Awodem,
does the gantry Display blank or having 2 orange dots?

Regards

Borhan

Log inor Register

to rate and post a comment

(49)
(20)

Awodem Alain

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 10, 2012 05:40

Dear Borhan,
the gantry display go blank with 3 green dots.

regards

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 12, 2012 05:54

Dear Awodem
i faced a similler trouble before and the solution was very strange
i removed the small 45 plug from the intercom board at the back of the ice box
I red that there is serial line go paralel with the network line
the source of the serial line is the digi board it could be bad

wish this help

Log inor Register

to rate and post a comment

(1)

Ivan Cip

re: Re: GE CT Lightspeed plus ( scanning hardware reset failed)

January 19, 2015 12:00

How can I fix the serial line?
I have a similar problem.

regards.

Log inor Register

to rate and post a comment

Deepak Roy Sharma

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 11, 2012 12:47

Hello Alain,

Please check Slip Ring Diss-connect status in Service Home page if it's showing high with values in Red then you have Dirty Slip Rings which needs to be cleaned. Also sometimes if we have any of DAS Converter board failed or other DAS Processing Board fails then also it makes OBC to stop communicating. Check both of these conditions and read error log in details thoroughly from start of problem start date/time. You have more clues on problem causes.

Deepak

Log inor Register

to rate and post a comment

(49)
(20)

Awodem Alain

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 11, 2012 09:47

Hello Deepak,
thanks for your reply
I check the slip ring and also cleaning it, but the problem still same, I also bypass the data connection from stc to obc it still same.
I took all board(CPU, LSCOM, Gentry I/O, KV control, HEMRC and MA control) from the obc and check it to another system, all board are ok.
I now thinking about Das convertor board how should I check it when I couldn't reset the scanning hardware? any process to do that?

Log inor Register

to rate and post a comment

Deepak Roy Sharma

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 12, 2012 08:03

Hello ALian,

please send me complete error log on my email id deepak@ritpl.com

Deepak

Log inor Register

to rate and post a comment

(34)
(8)

Ruben Cadima

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 13, 2012 05:58

Hello
Can you please send me the error log file, my mail is rubideo@gmail.com
Regards
Ruben

Log inor Register

to rate and post a comment

(34)
(8)

Ruben Cadima

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 13, 2012 05:57

Hello
Can you please send me the error log file, my mail is rubideo@gmail.com
Regards
Ruben

Log inor Register

to rate and post a comment

(9)

Clarence Knox

Re: GE CT Lightspeed plus ( scanning hardware reset failed)

March 13, 2012 11:26

You can perform a hard reboot. Shute the computer down then turn the power supply of or circuit breaker off, waite a few seconds then turn the circuit breaker back on then power up the computer. That helps some times.

Clarence Knox, BMET

Log inor Register

to rate and post a comment

Arjun Siva

re: GE CT Lightspeed plus ( scanning hardware reset failed)

August 28, 2020 10:27

How to check the obc

Log inor Register

to rate and post a comment

(2)

Hector Quintana

re: GE CT Lightspeed plus ( scanning hardware reset failed)

October 16, 2025 11:40

I had this problem once, to solve it i polished the slipring and realigned the carbons... Worked like a charm...

Log inor Register

to rate and post a comment