Over 150 Missouri Auctions End Today - Bid Now

GE CT MAX 640 problem

Mario Silva

GE CT MAX 640 problem

April 16, 2008 04:19

The equipment does not show up the leyend: "System preparation", and it is still in the black screen (never boot up). Could it be the DSPH or the ITP2 board? thanks

Log inor Register

to rate and post a comment

(53)
(2)
(1)

yasar demirel

Re: GE CT MAX 640 problem

April 16, 2008 05:19

at power on you may have an info whre the system stoped by watching leds on right side panel you have the tell this info first
of course if all power supplies are ok.

Log inor Register

to rate and post a comment

Mario Silva

Re: GE CT MAX 640 problem

April 17, 2008 12:02

ok, i watch the leds in the right panel. The leds nÂș 6, 5, 2 and 0 are lite up. How do you read the error code from this status leds?
thanks

Log inor Register

to rate and post a comment

(53)
(2)
(1)

yasar demirel

Re: GE CT MAX 640 problem

April 18, 2008 04:05

according to this information you have 32H message
(no 5,6 means 3 2 means 2 in binary)
a
so your DMC board or hard disk is defected
you can skip power on self test(i am not sure but you have check pls push down 8 dip switches )
we have full spare&books
regards

Log inor Register

to rate and post a comment

Mario Silva

Re: GE CT MAX 640 problem

April 18, 2008 10:52

ok, thanks for your help. I will try the DMC board, but, why the system does`t show the "system preparation"? this problem is independent from the DMC board, i think. any idea?

Log inor Register

to rate and post a comment

(74)
(5)
(10)

Tayfun Canli

Re: GE CT MAX 640 problem

April 19, 2008 02:00

Most probably you have a defective HDD. Please check that your HDD starts spin when you applied power and if it is ok try to reload software before order controller board, also some times the board which attached to HDD may fail..

Log inor Register

to rate and post a comment

(53)
(2)
(1)

yasar demirel

Re: GE CT MAX 640 problem

April 21, 2008 02:30

system does the self test first
if you want to skip the self test push 8 dip sitches down for that
so you may may have an idea
regards

Log inor Register

to rate and post a comment

Mario Silva

Re: GE CT MAX 640 problem

April 23, 2008 03:57

i had a new error on the status led, the leds 5 and 0 are lite on. This is the error 21H? i dont know if this is the correct error code. how do you read the leds to obtain the code?.

the screen never show up the system preparation leyend.

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

Re: GE CT MAX 640 problem

April 23, 2008 11:41

Mario,
If the machine is been down for a while
and you want to know what it is
send me an e mail y have experience on these systems.
Hablo espanol

Log inor Register

to rate and post a comment

(53)
(2)
(1)

yasar demirel

Re: GE CT MAX 640 problem

April 24, 2008 02:27

21H means MR4 board

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

Re: GE CT MAX 640 problem

April 17, 2008 11:00

Mario,
If there is no change in the leds while the system powers up, take a look at what leds are lite up in right side panel, that will give you an idea on where to start. and make sure you're booting from the hard drive, not from the floppy.

Log inor Register