ge ct nxi image problem

(172)
(26)
(1)

Marivic Mendoza

ge ct nxi image problem

July 03, 2012 11:38

we have a ct scan a GE NXI when we scan an image we are having a problem. the image in the inner part is dARK IN COLOR AND at the outer part is light in color...i think it is not an artifact but if you will judge the whole image it is not uniform in color...how can we correct this becaUse we are having a conformance test and we might fail in the uniform test.. thanks for the help

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: ge ct nxi image problem

July 04, 2012 06:43

Have you perform calibration? if yes, Perform a warm up error analysis, and check the channels that are outside tolerance.

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: ge ct nxi image problem

July 04, 2012 11:59

ok i will try this function and get back to you

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: ge ct nxi image problem

July 04, 2012 08:10

That is considered an artifacts. After you perform what mr.martinez suggested any updates now?

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: ge ct nxi image problem

July 04, 2012 08:10

I suggest please tell your service engr. to read the theory section of the service manual so that they will have an overview how each subsystems works. Likewise, do not spoon feed your service engr the specific things they should do from the suggestions you got from here. Give them a general overview and let them do the analysis and the specific things that should be done. That way your engr will learn a lot. The best way for the service engr to learn their machine is during installation and commissioning while the machine is not being endorsed. It will give your engr enough time to have hands on practice with less pressure from the buyer.

Always ask for them a detailed service report with detailed analysis what makes the systems failed.

Ask them questions how they come up with a solutions and what really went wrong everytime they solve a problem. if they can give you a well profound technical explanations then they learned and they will become one of the best troubleshooter in your area.

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: ge ct nxi image problem

July 04, 2012 11:59

thanks for your comment.. i will certainly do that

Log inor Register

to rate and post a comment

(19)
(8)

Dragos Mosneagu

re: ge ct nxi image problem

July 04, 2012 08:10

Can you send me some dicom images with your problem, at lest screen pic`s, in order to have an pertinent opinion?
My email is dragos.mosneagu@tmsro.ro

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: ge ct nxi image problem

July 04, 2012 11:59

i dont have any image but if you check the image which is in circle shape the inner part of it is dark in color as it goes to the outer side it will become light

Log inor Register

to rate and post a comment

(15)

EDMOND BUENAFLOR

re: re: re: ge ct nxi image problem

July 05, 2012 08:35

uniformity image problem... considered an artifacts.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: re: re: ge ct nxi image problem

July 05, 2012 08:35

If you are not an Engineer, my recommendation is for you to find one to work in your problem, and don’t risk failing the conformance test. The same way that you rely upon a radiologist for the diagnostic of your patients, you should use an engineer for the diagnostic of your CT.

Log inor Register

to rate and post a comment

(19)
(8)

Dragos Mosneagu

re: re: re: ge ct nxi image problem

July 05, 2012 08:36

Is it anatomical image, or you have only dark and light grey? If you have only grey try an retrorecon on some old rawdata if the images are ok the problem is on DAS, most probably the power supply be aware there are 4-5 power supply, check voltages and adjust it / replace if necessary.

1 person liked this
Log inor Register

to rate and post a comment

(15)

EDMOND BUENAFLOR

re: ge ct nxi image problem

July 04, 2012 11:59

i can check your problem if you want? you can contact me @09109411545

Log inor Register

to rate and post a comment