Hi. Do you have any solutions in my problem? Thanks in advance!
Host: OBC ERROR: 213501
Detector temperature error
The detector temperature, monitored by Gentry I/O, is out of tolerance.
Measured temp.: 22.64 C
Min. Level: 30.50 C Max. Level: 38 C
Marivic, Herewth Just an information what I knew are Input Voltage is 24 VoltDC, 60 Watt, temperature 35 Deg C, need 3 hours to stabilize, do you have service manual?, you can download at http://apps.gehealthcare.com/servlet/ClientServlet
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Marivic Mendoza
GE CT SCAN CT/I Detector temperature error
January 10, 2013 10:06
Hi. Do you have any solutions in my problem? Thanks in advance!
Host: OBC ERROR: 213501
Detector temperature error
The detector temperature, monitored by Gentry I/O, is out of tolerance.
Measured temp.: 22.64 C
Min. Level: 30.50 C Max. Level: 38 C
I already checked heater power supply 38V.
to rate and post a comment
Ronald
re: GE CT SCAN CT/I Detector temperature error
January 11, 2013 08:26
Marivic, Herewth Just an information what I knew are Input Voltage is 24 VoltDC, 60 Watt, temperature 35 Deg C, need 3 hours to stabilize, do you have service manual?, you can download at http://apps.gehealthcare.com/servlet/ClientServlet
to rate and post a comment
Marivic Mendoza
re: re: GE CT SCAN CT/I Detector temperature error
January 17, 2013 10:05
can you locate where is the 24voltsdc , 60 watt, 35 deg c?
to rate and post a comment
Dragos Mosneagu
re: GE CT SCAN CT/I Detector temperature error
January 11, 2013 08:46
Did you check continuity / resistance of the heater?
to rate and post a comment
Marivic Mendoza
re: re: GE CT SCAN CT/I Detector temperature error
January 17, 2013 10:04
where can i check the continuity/resistance of the heater and what should be the correct value
to rate and post a comment
xray
re: GE CT SCAN CT/I Detector temperature error
January 11, 2013 08:46
Check solid State Relay 46-188067P1
SS Relay can be found in the Laser Assembly near OBC