ge ct scan helical problem

January 07, 2013 11:37

we are encountering a problem with our ge ct scan ct/i. During Helical scan it will scan then it will stop scanner hardware. when you resume it will continue the scanning but it will stop scanner hardware again. the process is the same until it finish the whole scan. The error we are getting is:

ERROR: 253804
FEP DSP1:Hardware Communication Failure
FIFO ERROR: DAS FIFO Overrun Error
HDW Status: (0x400009e) (Data Type: 0x2)
View sequence number:7522
The FEP Hardware/software could not keep up with the SCOM Data

ERROR: 253914
FEP DSP1: Time out waiting for full dma list to complete.
save number:23
IIE: 0x26404001, DMA Control: 0xcc4004b, DMA count: 0x1

Can somebody help me figure out this problem..thanks in advance

Log inor Register

to rate and post a comment

Rajesh Nair

re: ge ct scan helical problem

January 09, 2013 09:09

Have you tried replacing the SCOM board? You can try cleaning up the slip rings too. Is the system giving any " Not able to write raw data" error?

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: ge ct scan helical problem

January 10, 2013 10:06

yes an error of "not able to write data sometimes appear

Log inor Register

to rate and post a comment

(5)

ashish

re: ge ct scan helical problem

January 09, 2013 09:10

I THINK THE DATA IS NOT REACHING FROM GANTRY TO OC. DURING HELICAL SCAN THERE IS CONTINUOUS DATA FLOW, MAY BE YOU ARE NOT FACING PROBLEM IN AXIAL AND STATIONARY SCAN. PROBABLE CAUSE IS DIRTY SLIP RING WHERE SIGNAL BRUSHES ARE MAKING CONTACTS. CLEAN THEM WITH HELP OF SAND PAPER.

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: ge ct scan helical problem

January 17, 2013 10:04

i already clean the slip ring but still stops during helical scan. do you have any idea where is the problem coming

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)