ge ct scan

(46)
(4)

marivic

ge ct scan

September 13, 2015 10:36

hi everyone,
we have a problem with our ge ct/i scan. it is an octane system. the issue is when we turn on the machine it will power on until it will display a hardware reset successful on the rx monitor but at the display monitor the message OC initializing still there. after awhile it shutdown the system until it will show the message-press restart to statt the system. the process will just repeat over and over. can someone tell me what causing the machine shutdown itself. i already tried to reload the software but only on the oc side but the problem still exist, thanks in advance

2 people liked this
Log inor Register

to rate and post a comment

(23)
(11)

Jerry Beeler

re: ge ct scan

September 14, 2015 08:30

System cannot recon rat gold test image so it shuts down. Could be bad Pegasus IG or high speed data disk or Motorola CPU board controlling Pegasus. Lights on Motorola bd and Pegasus board will give you clues as to if they are booting properly.

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 08:11

i check the IG board and the 4 leds is continous flashing.

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: ge ct scan

September 14, 2015 08:30

Hello! The problem is with your hardware of the scanner. What errors are displayed on the monitor during the starting up?

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 08:11

i dont see any error during bootup. i also turn off the switches in the table and gantry and turn on the console. it will show hardware reset fail but it will shutdown again itself

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: ge ct scan

September 14, 2015 08:30

Hi Angela.

It is very possible that the problem is that the raw data/scan disk is bad.
Try to stop the boot up process and log in as root/#bigguy and run reconfig with regeneration of database.

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 08:11

how will i do it?just click yes in regenerate data? i just want to ask if i do this will the calibration be lost? we already lost the system state of the machine i am if the calibration will be lost if we regenerate the data

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: re: re: ge ct scan

September 17, 2015 09:02

your calibration will be erased.

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: re: ge ct scan

September 15, 2015 06:19

Look at the /usr/g/service/log for the file ICE or icelog.
Send me an email to fito1971@yahoo.com to further assist you.

Log inor Register

to rate and post a comment

(23)
(7)
(14)

Giovanni Lombardo

re: ge ct scan

September 14, 2015 08:30

Change the keyboard or memory O2

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 08:12

when i run ide test for memory it say pass. regarding with the keyboard i have not replace it yet

Log inor Register

to rate and post a comment

(32)
(9)

HECTOR VELAZQUEZ

re: ge ct scan

September 14, 2015 08:30

It could be the recon system! Raw data disks, IG, ICE. Well... It is easy
to figure it out. Send me error log and i will help you hvelazquezm@gmail.com

Log inor Register

to rate and post a comment

(4)
(1)

Hakim Haffar

re: ge ct scan

September 14, 2015 08:30

Hi angela
Please check the ice box

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 11:04

can you tell me where is the ice box?

Log inor Register

to rate and post a comment

(29)
(11)
(16)

Vijay Sharma

re: ge ct scan

September 14, 2015 08:30

Please check if recon systems is booting successfully.

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 11:05

how will i know if the recon system is booting successfully

Log inor Register

to rate and post a comment

mahmoud abdelhaddy

re: ge ct scan

September 14, 2015 08:30

You have to try 1- the mouse connection and the keyboard
Check that and give to me the reply
On en.mahmoud.abdelhady@gmail.con

Log inor Register

to rate and post a comment

(3)

kumar sujeet

re: ge ct scan

September 15, 2015 08:09

Facing the same problem dear,
Please help if you figured out your problem

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: ge ct scan

September 15, 2015 11:05

have you tried the suggestions given in this thread?

Log inor Register

to rate and post a comment

(23)
(11)

Jerry Beeler

re: ge ct scan

September 15, 2015 11:05

Your problem is not the mouse and or the keyboard. If you regen the high speed data disk no system info is lost it only reformists the raw data disk which has nothing but raw scan data on it. Your problem is in the scan recon system somewhere. Which does include the scan data disk. You can (should) watch the screen as the scan disk is regenerated to make sure there are no errors.

Log inor Register

to rate and post a comment

(32)
(9)

HECTOR VELAZQUEZ

re: ge ct scan

September 17, 2015 04:20

Does some body realized than this angela or marivic has been changing his her name? Well... I don't trust
In people like her him

Log inor Register

to rate and post a comment