Hello
I have problem in ge ct system model lemage suprem there artifact in image and when did air calibration give me SD error
And there serculer in image I hope to help me to solve it
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Ge ct-scanner
December 03, 2015 01:54
Hello
I have problem in ge ct system model lemage suprem there artifact in image and when did air calibration give me SD error
And there serculer in image I hope to help me to solve it
to rate and post a comment
Joseph Jenkins
re: Ge ct-scanner
December 04, 2015 09:03
The compensator inside the collimator is probably cracked and needs to be replace
to rate and post a comment
Aleksander Proof
re: Ge ct-scanner
December 04, 2015 09:03
Hi!
If you have circular artifacts in image, I think your problem in detectors. You must check it.
to rate and post a comment
Abdelmuin Khatib
re: re: Ge ct-scanner
December 07, 2015 04:29
Hello
How I can test it?
to rate and post a comment
Wang Zhixuan
re: Ge ct-scanner
December 04, 2015 09:03
You need to check DAS and detector, could you check them by yourself?
My email ID is 362689810@qq.com
to rate and post a comment
DHINESH RAJ
re: Ge ct-scanner
December 04, 2015 09:04
Its a solid state detector or xeonon detector? First check das power supply....
Its ok do cam linierty test....
to rate and post a comment
Abdelmuin Khatib
re: re: Ge ct-scanner
December 07, 2015 04:29
Hello
Das power ok, if linierty not correct what I must do?
to rate and post a comment
Abdelmuin Khatib
re: re: Ge ct-scanner
December 09, 2015 08:49
Hello
It is solid state detector
I did linierty test it not good
to rate and post a comment
Joseph Jenkins
re: Ge ct-scanner
December 08, 2015 08:06
Check the compensator in the collimator.... They all crack eventually and cause ring artifact.... Will
Be a hairline fracture running though it ...