Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Ge CTe device 4 error

(47)
(12)

Jose Augusto

Ge CTe device 4 error

January 25, 2016 09:59

Hy,

I'm able to "run diagnostic" and all are passed except "SCSI test" with this error message :

Device 4 failed DMA test.
error code =IP32.08.0c.01.04,W1=00000001, W2=00000010

Can`t load media to instal SW.

Any body help me?

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: Ge CTe device 4 error

January 26, 2016 09:43

Are you have O2 computer? If yes i can help you. Email me ermsismed@gmail.com

Log inor Register

to rate and post a comment

(51)
(24)
(3)

Gustavo Gnecco

re: Ge CTe device 4 error

January 26, 2016 09:44

Hi Junior,

I have seen the photos on the whatsapp you sent me.

To run this diagnostic, disconnect all the devices from SCSI and go testing on by one. Probably your problems is with the Serial Exp. Start with this.
You told me you have another machine to take parts, try this part first.

The other problem you sent me the photos, maybe you have another bad disk. Was The format successful ??

Let me know.

Gustavo.
www.emsts.com.br

Log inor Register

to rate and post a comment

stats
DOTmed text ad

YAG Laser Service Specialists

Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)