Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

GE CT/E Problem

(7)

hassen ameur

GE CT/E Problem

December 09, 2007 11:46

Hello everybody

We have a ct/e problem. Booting is okay but when we try tube warm up or daily calibration we get the following errors:
Gantry error code 13-1002-01 and
XG error phase is 3 code 40 40-0106H
We checked the cables between rotation board and tube.. but nothing
The rotation board leds are okay
Could it be the tube (stator fault)? because it has been arcing for a while now.

Can anybody help

Hassen

Log inor Register

to rate and post a comment

Ralph Childs

Re: GE CT/E Problem

December 11, 2007 03:18

Hassen,

Much like my reply to the CT/I error listing last week.

Save time... Call your local tube supplier.

Scooter