Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE CTe XG error phase is 4 code is 60

Tags: CT, GE

(47)
(12)

Jose Augusto

GE CTe XG error phase is 4 code is 60

February 21, 2013 03:23

Hello all,

I have a problem with GE CTe.
When i make a exposure, shows:

Scout scan error code:20-1001-43, and XG error phase is 4 code is 60. Deail error code: 60-0313H Expousure error.

Someone help-me
Thanks. Jose.
terxlu@gmail.com

Log inor Register

to rate and post a comment

Mario Wug

re: GE CTe XG error phase is 4 code is 60

February 21, 2013 10:43

Run the no load KV test on system to make sure its the tube. Usually that error can be caused by the HV tank.

Log inor Register

to rate and post a comment

(51)
(24)
(3)

Gustavo Gnecco

re: GE CTe XG error phase is 4 code is 60

February 21, 2013 10:43

Hi Jose Augusto,

Probably your tube is gone. 90% chance
If not, the problem is with other part in the Generator like HV Tank or KV Control

We can test your generator and provide the tube or any other parts of this generator

Call me

******************************************************

Olá José Augusto,

Provavelmente o tubo pifou. 90% de chance
Se não for o tubo, o problema esta em outra parte do gerador como o Tanque de Alta Tensão ou a placa de controle do KV.

Nós podemos testar seu gerador para ter certeza da peça defeituosa, assim como possuímos todas as peças desses gerador em estoque

Entre em contato para solucionarmos seu problema

Grato

Gustavo.
www.emsts.com.br

1 person liked this
Log inor Register

to rate and post a comment

(31)
(1)
(12)

S Goyal

re: GE CTe XG error phase is 4 code is 60

February 21, 2013 10:43

The most probable cause for this error is tube. The second possibility is HV tank. The third possibility is KV control board. If tube is old then check by changing the tube.