Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

GE CT/e

Tags: CT, GE

(10)

Ashu

GE CT/e

January 11, 2013 08:45

Dear friends, .

After switching On, GE CT/e, we noticed error " System found back dated. All FlexTrial Options are uninstalled. User can no longer use these options".
The system booted up normally after that but the system date was backdated.
I think this error error was due to CMOS battery. Please adv. whether we need to change the battery. If yes, please adv,. whether the date needs to be updated only in date settings or in BIOS as well.
Thanks in adv.

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: GE CT/e

January 12, 2013 11:23

Hello, Ashu! You need to replace a battery on the computer. The battery type is CR2032. It is simple the computer battery. After replacing you should set the current date and time.

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: GE CT/e

January 12, 2013 11:23

You can change the date in BIOS or in system. There is no difference

Log inor Register

to rate and post a comment

(10)

Ashu

re: re: GE CT/e

January 14, 2013 12:35

Thanks Dmitry for your reply. I really appreciate your help.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: GE CT/e

January 12, 2013 03:46

If your system is IRIX based, you can do it clicking Reconfigure

Log inor Register

to rate and post a comment

stats