GE CTi CT Scanner booting problem

(58)
(6)

Elias Valdes

GE CTi CT Scanner booting problem

December 12, 2010 05:19

Hello:
I have a booting problem with a GE CTi CT scanner. When starting the system most of the time towards the end of loading the software, it doesn't performs the scanning reset, RECON is paused and the system sends a "PANIC: tblmiss: invalid kptblentry" error message. I have loaded software, replaced the SBC Software and Scan Data Disks. After loading the software I was able to even perform several reboots without any problems, I was able to perform scans and get images without any problems, but after turning off the scanner, problems started again. I have swaped the memory banks in the Indigo2 and I have even tried to work with just one memory bank and I get the same error message. I will appreciate any suggestions to solve the problem. Thanks
Elias

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: GE CTi CT Scanner booting problem

December 13, 2010 09:22

i can help you ,please contact me.

Log inor Register

to rate and post a comment

Wade Nakatani

Re: GE CTi CT Scanner booting problem

August 04, 2011 10:12

saw the email about the person having a problem with sbc board not rebooting. He replaced the sbc board and power supply with no solution to his problem. You say you have the solution....we are having same problem...please help. Thanks, Wade
My email address is mistyrays@yahoo.com Thank you so much

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: GE CTi CT Scanner booting problem

December 14, 2010 09:14

I have never seen that particular problem, but given what you have replaced so far, I would check the power supply to the SBC rack and suspect the SBC computer board.

Log inor Register

to rate and post a comment

(58)
(6)

Elias Valdes

Re: GE CTi CT Scanner booting problem

December 16, 2010 09:34

I have already replaced both.

Log inor Register

to rate and post a comment

(67)
(3)

farhan zeb

Re: GE CTi CT Scanner booting problem

September 17, 2011 07:12

INDIGO2 4.4 version


1_) dear friend i have loaded the software first i loaded the OS then

Applications and after that i have to configure or load the SBC with OC when i type the command loadSBC so the

following error occurs Not enough free spcae can not load network software

when i try to connect to SBC i type the command cu sbc but following error occured connect failed: cant't access

device


i have tried all the procedure from LFC but still same problem i have tried to bring it to the MC6080> prompt but

still nothing happened still same problem






but when i use the command rsh sbc and reconfig the sbc and after st so during initialization system goes shut down

itself .



NOTE:


when i have loaded the application software so after that when i open the unix shell so an message

appeared executing install.dasm no dasm's found



so please help me i'll wait for your reply





2_) Dear friend i have lost the system state also of cti indigo2 system and and the oc disk drive have been crashed

i have

replace the disk drive and loaded the software from cold so i have lost my system state so now i have loaded the

software but i have no system state but there is need of the system state so what should i do now there is the

method to restore files in the LFC but how should i access the unix shell with crash disk drive so please tell me

what will be the alternative way to skip this step and start the applications.




Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)