August 29, 2012 10:59

hello there. i have a problem with my ct scan.during filming the ct scan system hangs. it means when i do the planning for printing the images the system is hanging or not responding.when i check the error it says..DCM low level error.system error(502):network is unreachable(128).can you help me for this



Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

re: ge cti hispeed

August 31, 2012 10:21

Disconnect network cable going towards dicom workstation and also Medical imager and then check but disconnect one at a time so that you may isolate your problem if not then email me at tomcharles41@yahoo.com