This GE cti is a new installation with a PDU number 327106G1 and an Octane console.During power up sequence The A3K3 relay does not engage to give power to the console.
Everything has been wired properly. Do i miss to make some jumper?
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
bens prado
GE cti power up problem
July 21, 2011 02:53
This GE cti is a new installation with a PDU number 327106G1 and an Octane console.During power up sequence The A3K3 relay does not engage to give power to the console.
Everything has been wired properly. Do i miss to make some jumper?
Any suggestion?
thanks
to rate and post a comment
Manuel Casanova
Re: GE cti power up problem
July 26, 2011 02:59
CHECK THE INPUT SETTING VOLTAJE ON THE TRANSFORMER PRIMARY.
to rate and post a comment
bens prado
Re: GE cti power up problem
July 27, 2011 08:48
ok thank you. we can power up now
to rate and post a comment
salman khan
Re: GE cti loadSBC problem
August 20, 2011 06:22
hi ihave problem when i type during lfc procedure loadSBC following erroe occurs what should i do
not enough space to continue
to rate and post a comment
Kim Schofield
Re: GE cti loadSBC problem
August 21, 2011 11:09
You probably have a bad disc drive
to rate and post a comment