Hello,
I have a problem with this machine
When i start the procedure, shows:
"Scanner hardware stopped scan"
Someone help me?

Thanks,
Jose

Log inor Register

to rate and post a comment

(20)
(2)

Anatoliy Mamykin

re: GE CTi Scanner error

July 09, 2015 09:06

Hi,
i will try to help you.
Please send the log to mamykin.t@gmail.com.

2 people liked this
Log inor Register

to rate and post a comment

(47)
(12)

Jose Augusto

re: re: GE CTi Scanner error

July 12, 2015 02:18

Hi,
I found a bad R13 (10 ohms x 20W) on PDU control board.
Replace and work!
5 days ok!
You know what happened?
Any reason for R13 open?

Thanks for you help