TOM CHARLES

GE cti

April 10, 2012 07:21

Sir today when we performed a scan of axial at 120kv and 200mA so an error appeared scanner hardware stopped scan what should be the reason and when i saw the error log so the error was what should i do further to solve this problem and second problem is that when i make a scan on axial it contains ring artifacts solution to this will also be appreciated





CT01 SBC0 scanmgr
do_axial_complete.m 1.17 do_axial_complete() 586

Function: Data Acquisition : SBC Processing
Scan: 13/2/1 Type: Axial Scan Seq: 22001
Axial Scan Status Message: The System Aborted Scanning:, views expected 984 - views collected 154

EN 324


SR 338
1333908135 2 1 Sun Apr 8 18:02:15 2012 183109 2
CT01 OBC Generator
gen_abort.c 1.33 No Method 248

Function: XRAY GENERATION : KV Loop
Scan: 13/2/1 Type: Axial
Exception Level: Pri/Most Severe Time: 18:02:15:305 Log Series: 98
KV fault detected: Anode overcurrent.
Fault latch address = ffcff3 hex, Bit = D4

EN 338


SR 510
1333908135 3 1 Sun Apr 8 18:02:15 2012 183296 2
CT01 OBC Generator
gen_abort.c 1.33 No Method 654

Function: No System Function Reported
Scan: 13/2/1 Type: Axial
Exception Level: Sec/Soft Time: 18:02:15:330 Log Series: 98
Selected Technic: 120 kV, 200 mA, 0 % Mod, Backup Timer 2162.
VarmA: (Index: Start 0, Skip 0, Last 0) Last Cmd 200 mA.
Monitoring: (Measured 197.4 mA vs Ref 200 mA).
3 Total Samples, Tube Spit Ctr 7.
(Rotor 0x16008204, Gen 0x8c1995ff, Mon 0x00000000)

EN 510


SR 268
1333908150 0 1 Sun Apr 8 18:02:30 2012 245043 4
CT01 SBC0 scanmgr
ScanMgrEventHandler.m 1.35 processUnsolicitedLetters: 714

Function: Data Acquisition : SBC Processing
Scan: 13/2/1 Type: Axial Scan Seq: 35001
Scanning System Aborted Scanning

EN 268


SR 295
1333908151 0 1 Sun Apr 8 18:02:31 2012 183257 2
CT01 OBC Control
gen_abort.c 1.33 No Method 716

Function: XRAY GENERATION : KV Loop
Scan: 13/2/0 Type: None/Unknown
Exception Level: Pri/Soft Time: 18:02:30:330 Log Series: 100
KV board tube spit counter = 10

EN 295


SR 324
1333908151 1 1 Sun Apr 8 18:02:31 2012 245071 4
CT01 SBC0 scanmgr
do_axial_complete.m 1.17 do_axial_complete() 586

Function: Data Acquisition : SBC Processing
Scan: 13/2/1 Type: Axial Scan Seq: 35001
Axial Scan Status Message: The System Aborted Scanning:, views expected 984 - views collected 150

EN 324


SR 339
1333908151 2 1 Sun Apr 8 18:02:31 2012 183109 2
CT01 OBC Generator
gen_abort.c 1.33 No Method 248

Function: XRAY GENERATION : KV Loop
Scan: 13/2/1 Type: Axial
Exception Level: Pri/Most Severe Time: 18:02:30:335 Log Series: 101
KV fault detected: Anode overcurrent.
Fault latch address = ffcff3 hex, Bit = D4

EN 339


SR 512
1333908151 3 1 Sun Apr 8 18:02:31 2012 183296 2
CT01 OBC Generator
gen_abort.c 1.33 No Method 654

Function: No System Function Reported
Scan: 13/2/1 Type: Axial
Exception Level: Sec/Soft Time: 18:02:30:360 Log Series: 101
Selected Technic: 120 kV, 200 mA, 0 % Mod, Backup Timer 2162.
VarmA: (Index: Start 0, Skip 0, Last 0) Last Cmd 200 mA.
Monitoring: (Measured 196.9 mA vs Ref 200 mA).
4 Total Samples, Tube Spit Ctr 10.
(Rotor 0x16008204, Gen 0x9c1995ff, Mon 0x00000000)

EN 512


SR 334
1333908190 0 1 Sun Apr 8 18:03:10 2012 184142 4
CT01 STC Scan Control
scTrigSeqActions.c 1.68 No Method 928

Function: No System Function Reported
Scan: 13/2/2 Type: Axial
Exception Level: Pri/Soft Time: 18:03:10:990 Log Series: 103
ISD missed.
Rx ISD: 1000 mS
Actual ISD: 3310
More Data Follows:

EN 334


SR 534
1333908190 1 1 Sun Apr 8 18:03:10 2012 184188 4
CT01 STC Scan Control
scTrigSeqActions.c 1.68 No Method 933

Function: No System Function Reported
Scan: 13/2/2 Type: Axial
Exception Level: Pri/Soft Time: 18:03:10:995 Log Series: 104
Last Hdw Rdys- Offsets: 0x400 CPB: 0x10 X-ray: 0x80
Timing (in mS) since end of last scan.
Rx rcvd: -1955 FEP/AP Offset Rdy: 215 Offsets start: 220
CPB Enable: 3155 Offsets stop: 3170 Cradle Rdy: 805
FEP/AP View Rdy: 3205 Trig Hdw Enable: 3215 Exp Enable: 3300

EN 534

Log inor Register

to rate and post a comment

(50)
(5)

ricardo Usandivaras

Re: GE cti

April 12, 2012 08:31

Check the power supplay of the sensor,one time it happen to me,was 1 capacitor blowed,check 5 volts

Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

Re: GE cti

April 13, 2012 08:51

which one sensor and which supply

Log inor Register

to rate and post a comment

(50)
(5)

ricardo Usandivaras

Re: GE cti

April 14, 2012 02:26

The rx detector power supply

Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

re: Re: GE cti

April 25, 2012 11:43

Sir i can you please locate me the section in manual that which one power supply it is and another thing is that during scan scanning hardware stoped scan message appears and then in error log this kind of error appears i have posted anode over curren etc i have replaced HV tanks but still same problem

Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

re: Re: GE cti

April 25, 2012 11:43

send me your email address i will send you the complete detail

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)