G.E. DMR E01(006/002) E05(079/005) Help me

Tags: Mammo, GE

(1)

julio martinez

G.E. DMR E01(006/002) E05(079/005) Help me

August 07, 2015 06:42

Could someone help me with the resolution of errors showing DMR + mammography, turning on the mammography, the test passes on and is ready for shooting of x rays in any manual or automatic mode, a test ray X this is interrupted and displays E01 (006/002) and E05 (079/005) errors, need help to solve them, following tests have changed the high voltage tank, 701 card, the card 400PL1, 400PL2 and 400PL3, cable high voltage x-ray tube, card 300PL1.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: G.E. DMR E01(006/002) E05(079/005) Help me

August 10, 2015 11:44

Generator: signal Kv_inf_5 (less than 5 kv) not rising,