Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE DMR MAMMO problem

Tags: Mammo

(27)
(6)

Luis Vargas Carrillo

GE DMR MAMMO problem

October 08, 2015 07:03

Hello,

I replaced the gantri batt and I want to calibrate it, so I changed the dip switch in the generator and turned the MAMMO on but nothing happends in the options menu it just says medical and language options, what am I doing wrong?

Thanks!

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: GE DMR MAMMO problem

October 09, 2015 05:59

did you enter the correct password email me
**redacted**

Log inor Register

to rate and post a comment

(7)
(1)

vijay bhatti

re: GE DMR MAMMO problem

October 10, 2015 09:57

Dear friend,you can't enter the install menu by entering the password. You need support of trained engg. Who have the software to connect equipment with laptop. you also contact me on 00919814803444
Thanks,
vijay.bhatti@yahoo.com

Log inor Register

to rate and post a comment

(15)
(4)

DEVENDRAN DRAN

re:GE DMR MAMMO problem

October 10, 2015 09:58

devendeanraja5679@gmail. com

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: GE DMR MAMMO problem

October 10, 2015 09:58

Do not do the same mistake by changing the battery in the generator. You must have a backup floppy first. A mammocom software and a dos computer with serial port also.
You must change the dip switch position when the system is on and every time you switch the syst on.

Log inor Register

to rate and post a comment

(6)

Jose Luis Baltasar

re: GE DMR MAMMO problem

October 10, 2015 10:00

I can help you, please send me a e-mil jose53fuenla@gmail.com

Log inor Register

to rate and post a comment

(16)
(5)

Andrzej Kunz

re: GE DMR MAMMO problem

October 11, 2015 09:27

Now you should upload the software from a floppy disk factory using stat.exe via RS232 (mammocom). If you upload software correctly you get the opportunity to enter to system from console. You should now 4 digits password. If you did early the backup, you can upload it using the stav.exe. In this way, you restore the last system settings.

Good luck.

Andrzej

1 person liked this
Log inor Register

to rate and post a comment

(4)

azad singh dahiya

re: GE DMR MAMMO problem

October 13, 2015 09:16

Dear sir, if you want to enter in installation menu then you should change the dip switch in gantry CPU board also so that both become same,then only you can enter.

Log inor Register

to rate and post a comment

(4)

azad singh dahiya

re: GE DMR MAMMO problem

October 13, 2015 09:16

Sorry,not gantry but generator CPU board dip switch you have to change for entering installation menu

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: GE DMR MAMMO problem

October 13, 2015 11:15

That is a very important point. You must change position when generator is ON!!!

Log inor Register