Hi I keep getting error code E003 rotor failure 017/004 can someone give me a hand on this matter I have changed just about every board that is realted to this issue.

Thank you very much for your help

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: GE DMR

April 20, 2015 10:07

Is the stator making noises? do you get the error @ CC and MLO?

Log inor Register

to rate and post a comment

ErnestoQ

re: GE DMR

April 20, 2015 10:14

Did you ohm out the rotor(stator) cabling to check for continuity?

Log inor Register

to rate and post a comment

(54)
(7)
(9)

Deborah Breland

re: GE DMR

April 20, 2015 12:08

Dear Sir, can you please let me know which boards you have changed? Feel free to email me at debi.breland@digitecmedical.com and we will help get your issue taken care of.

Log inor Register

to rate and post a comment

(19)
(4)

walter molina

re: GE DMR

April 20, 2015 12:09

Hello can you check the rotor board in the generator unit , and check all fuses , or the xray tube its bad.good luck .