GE Essential (DS) Gantry CPU problem

Tags: Mammo, GE

Posted by an Anonymous User

GE Essential (DS) Gantry CPU problem

May 07, 2024 10:11

Hello. There is problem with Gantry CPU of GE essential DS. Timekeeper battery was dead and now there isn't LAN connection with this device--No ping on 192.168.1.40. Is it possible to reconfigure CPU board?
Thanks in advance

Log inor Register

to rate and post a comment

(18)

Annas

re: GE Essential (DS) Gantry CPU problem

May 08, 2024 12:06

No only replace is the option , if you remove battery than only option is replace the gantry cpu board.

Log inor Register

to rate and post a comment

(28)

Giorgio66

re: re: GE Essential (DS) Gantry CPU problem

May 10, 2024 10:52

Even if I had Mammocom SW and DMR (GEN+Gantry) backup files? Thank you very much.

Log inor Register

to rate and post a comment

(18)

Annas

re: re: re: GE Essential (DS) Gantry CPU problem

May 15, 2024 12:32

yes till you connect its ip
192.168.1.40

Log inor Register

to rate and post a comment