Dear all
I have a problem with ge Flourostar 7900 .
When power on ,booting is complete, no errors.
At auto flouro mode ,no image until kv reached 110kv
and the image soft with no details.
Thanks for any advance.
Put in manual mode 70kv 3 ma , check your kv output with scope or non-invasive tool, if you dont have sxact kv your monotube XRT is dead, most probably.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
a.h.m
Ge Flourostar 7900 problem
March 14, 2025 10:38
Dear all
I have a problem with ge Flourostar 7900 .
When power on ,booting is complete, no errors.
At auto flouro mode ,no image until kv reached 110kv
and the image soft with no details.
Thanks for any advance.
to rate and post a comment
Carlos Pons
re: Ge Flourostar 7900 problem
March 17, 2025 10:02
Put in manual mode 70kv 3 ma , check your kv output with scope or non-invasive tool, if you dont have sxact kv your monotube XRT is dead, most probably.
to rate and post a comment
Holger Fritze
re: Ge Flourostar 7900 problem
May 05, 2025 02:12
check small focus, most its interrupted. If so, please email me.
to rate and post a comment
Holger Fritze
re: Ge Flourostar 7900 problem
June 09, 2025 10:10
did you solve? what was the cause?
to rate and post a comment