We are working on GE FXi. We changed the tubes and
After closing the Gantry covers very well,The CT Monitor says "Safety switch open".
What can we do to fix this?

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: GE FXi COVER PROBLEM

August 21, 2012 06:41

look for the cover detector switch, located at vertical sides of the gantry.
probably you dont fit correctly the plates that push the small switches

Log inor Register

to rate and post a comment

(30)
(9)

Dean Kenney

re: GE FXi COVER PROBLEM

August 21, 2012 06:41

If you just took off the front cover then its the safety switch on the right side front cover knee high not good contact....also rt side cover behind the 1ft front by 1ft cover is another safety switch check that...back cover facing table left side knee high thats all folks