Our positioning lights on the scanner could not come on and the CT refused to scan. Error message is :TGP TASK TIME OUT. Then it displays XG ERROR.
Any help is most appreciated.

Log inor Register

to rate and post a comment

(28)

Veronica Alvarez

Re: GE FXi ERROR CODE

July 25, 2011 03:51

your generator is detecting a problem with x-ray need to send the error that has after the xg error to determine if you have a low ma over ma, over kv or under kv problem, ir the rotor is turning, can you hear the tube prepare to do x-ray, can you hear the generator do some thing,???? please advise mor information or call us to 3033042719

Log inor Register

to rate and post a comment

(31)
(12)
(2)

Thomas Onyema

Re: GE FXi ERROR CODE

July 25, 2011 05:16

Thanks Veronica.The error: 1)GANTRY ERROR CODE: 13-1002-17. Error detected by TGP(GP)driving power of servo amp is out of specification.
2)GANTRY ERROR CODE:13-1002-04 .Error detected by : TGP servo amp alarm is detected.

Log inor Register

to rate and post a comment

(13)
(6)
(5)

Nirvikar Yadav

Re: GE FXi ERROR CODE

July 25, 2011 07:22

Most probably Servo Amp is faulty. Contact me on +61449552069.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: GE FXi ERROR CODE

August 07, 2011 08:32

You need to be clear, if the error is from XG or SVAmp.

This a very different stages of failure

Log inor Register

to rate and post a comment

(42)
(4)

salman khan

re: Re: GE FXi ERROR CODE

January 28, 2013 04:49

Exactly Servo Amp is at fault