i have a indigo system for my ct/i ct scan.when i power up the machine it will load the application but after some time a message will appear a dumping error..low memory..something like that..when i restart the system it will still show same problem..i already cahnge the memory module and the cpu but no luck at all.can anyone help me figure out this problem
Can you send SYSLOG log file in /var/adm folder which will give details of actual computer hardware problem inside. Software load will not help if we have some Indigo Hardware problem which needs to be doagnosed first.
If you can open a shell, run storelog. From what you are describing, it sounds like your disk is full and applications cannot start. Running storelog may free up enough disk space to allow you to run applications and delete some patient images
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Marivic Mendoza
ge hispeed ct/i
June 22, 2012 09:24
i have a indigo system for my ct/i ct scan.when i power up the machine it will load the application but after some time a message will appear a dumping error..low memory..something like that..when i restart the system it will still show same problem..i already cahnge the memory module and the cpu but no luck at all.can anyone help me figure out this problem
to rate and post a comment
Juan de la Cruz Martinez
re: ge hispeed ct/i
June 22, 2012 09:27
Reload software. What is call Load from Cold.
to rate and post a comment
Deepak Roy Sharma
re: re: ge hispeed ct/i
June 23, 2012 01:57
Hello Marivic,
Can you send SYSLOG log file in /var/adm folder which will give details of actual computer hardware problem inside. Software load will not help if we have some Indigo Hardware problem which needs to be doagnosed first.
Deepak
to rate and post a comment
Marivic Mendoza
re: re: re: ge hispeed ct/i
June 24, 2012 11:53
i cannot go to the folder /var/adm because it will just show the dumping error after it finnish the loading the application
to rate and post a comment
Elliot Marcus
re: ge hispeed ct/i
June 23, 2012 01:57
If you can open a shell, run storelog. From what you are describing, it sounds like your disk is full and applications cannot start. Running storelog may free up enough disk space to allow you to run applications and delete some patient images
to rate and post a comment
Marivic Mendoza
re: re: ge hispeed ct/i
June 24, 2012 11:53
it cannot run the storelog because i cannot load the application
to rate and post a comment
Joginder Singh
re: ge hispeed ct/i
September 29, 2015 05:52
I have CTI with indigo. Its shows your current. Drive is less than 4gb. Software cannot be loaded. Pls help I have HDD with 9gb space