Victor Lima
Ge HiSpeed Error 70
November 24, 2012 08:50
Hi, When turn on the CT in the current message: XG error phase 2 code is 70. Detail error code 70: - 0503H Power Supply error: Invert Gate Power Supply error. Is it in brushes or what power supply?? Thanks a lot,
to rate and post a comment
TOM CHARLES
re: Ge HiSpeed Error 70
November 25, 2012 04:31
Hello friend Run HV Power power diagnostics and see for the error It must be also Mains , Cable Pb , gate command board (inverter) or Kv control check these sections also
Gustavo Gnecco
Hi Victor, Forget the brushes and power supply. Your problem is in the Power Unit. You have to run the HV power diagnostics, but this have to be done by a trained and experienced person Call us, we can help you solve your problem Regards Gustavo. www.emsts.com.br
Wang Zhixuan
November 25, 2012 08:39
The problem can be solved by us.Please tell me where is the Hispeed CT scanner in.My email is hzwzx@yahoo.com.cn.
Carlos Duran
November 26, 2012 10:21
Is the power that feeds the IGBTs, at control side. If this power is not present, the inverter can not operate, even the power is present at slip ring
xray
November 26, 2012 10:24
Hi, Potential cause: 1. Mains 2. Cable pb 3. Gate command board (Inverter) 4. kV control Recommended action/Troubleshooting guide: Run HV power diagnostics BR
Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com
Victor Lima
Ge HiSpeed Error 70
November 24, 2012 08:50
Hi,
When turn on the CT in the current message:
XG error phase 2 code is 70. Detail error code 70: - 0503H Power Supply error: Invert Gate Power Supply error.
Is it in brushes or what power supply??
Thanks a lot,
to rate and post a comment
TOM CHARLES
re: Ge HiSpeed Error 70
November 25, 2012 04:31
Hello friend
Run HV Power power diagnostics and see for the error
It must be also Mains , Cable Pb , gate command board (inverter) or Kv control check these sections also
to rate and post a comment
Gustavo Gnecco
re: Ge HiSpeed Error 70
November 25, 2012 04:31
Hi Victor,
Forget the brushes and power supply. Your problem is in the Power Unit. You have to run the HV power diagnostics, but this have to be done by a trained and experienced person
Call us, we can help you solve your problem
Regards
Gustavo.
www.emsts.com.br
to rate and post a comment
Wang Zhixuan
re: Ge HiSpeed Error 70
November 25, 2012 08:39
The problem can be solved by us.Please tell me where is the Hispeed CT scanner in.My email is hzwzx@yahoo.com.cn.
to rate and post a comment
Carlos Duran
re: Ge HiSpeed Error 70
November 26, 2012 10:21
Is the power that feeds the IGBTs, at control side. If this power is not present, the inverter can not operate, even the power is present at slip ring
to rate and post a comment
xray
re: Ge HiSpeed Error 70
November 26, 2012 10:24
Hi,
Potential cause:
1. Mains
2. Cable pb
3. Gate command board (Inverter)
4. kV control
Recommended action/Troubleshooting guide: Run HV power diagnostics
BR
to rate and post a comment