bens prado

GE hispeed error

September 15, 2012 11:11

Good day,

A GE HiSpeed advantage ct scan report an error messages during boot up. " scan hardware reset failed" at the error log it posted " 38volts
power supply for the stepper motor is too low (35V)"

May i know what could be the possible problem and solution?

Thank you

Log inor Register

to rate and post a comment

(43)

(2)

Ernesto Guerra

re: GE hispeed error

September 15, 2012 06:10

It could be few things one is the 38v power supply, or the collimator power module. Check your fuses, power supplies.............................................

Log inor Register

to rate and post a comment

(30)
(9)

Dean Kenney

re: GE hispeed error

September 15, 2012 06:10

Collimator

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: GE hispeed error

September 15, 2012 06:10

the collimator PS

Log inor Register

to rate and post a comment

(36)
(6)
(4)

Galal Hamood

re: GE hispeed error

September 16, 2012 11:22

from the error message , it seems that the motor is not working proberly to get back to home position .

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: GE hispeed error

September 16, 2012 07:59

Disconnect Collimator Servo Motor Drive And check the voltage on its Power Supply output. If missing, repair the P/S. It is a very simple analogue P/S with few components on it. If Voltage present on output that time (without load) it is maybe just a few adjustment on the P/S and becomes operational. The 38VDC is not a criteria on collimator P/S. You can lower it and then the operating Current will be lower so, the P/S can restart. I faced similar problem before. Good luck

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: GE hispeed error

September 16, 2012 10:43

Good day .

Thank you so much i will do what you have all suggested.

Again thank you

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: re: GE hispeed error

September 21, 2012 08:22

I found bad collimator servo db drive.
thank you to all

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)