GE HiSpeed FX/i problem

Alexander Tecson

GE HiSpeed FX/i problem

May 22, 2012 09:26

Good day,
I am having problem in Ge HiSpeed FX/i, the system failed to performed air calibration and with error in message log:

File:initair_calc.c
SD:status=1, error channel =10

File:initair_calc.c
Aperture Check Error 0 degree

File:initair_calc.c
Aperture Check Error 90 degree

File:initair_calc.c
Aperture Check Error 180 degree

File:initair_calc.c
Aperture Check Error 270 degree

File:initair_calc.c
InitAir:status =1

Any help and suggestion,

Thanks in advance.

Log inor Register

to rate and post a comment

Mark Parris

re: GE HiSpeed FX/i problem

May 22, 2012 06:01

sounds like a cracked compenstaor or bowtie filter

1 person liked this
Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE HiSpeed FX/i problem

May 23, 2012 08:58

thanks for the reply,. I already check the bowtie filter but no cracks present in the filter.

Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE HiSpeed FX/i problem

May 24, 2012 11:47

thanks for the reply, found defective cam board.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: GE HiSpeed FX/i problem

May 22, 2012 07:27

Try to do service scan, then check error message box. If you get aperture error again, means collimator problem.
If there is no error, try to do restorestate, to fix cal tables.

1 person liked this
Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE HiSpeed FX/i problem

May 23, 2012 11:01

Thank you for the reply,.

In service scan, no error appeared/occurred both stationary(0,90,180,270) and rotating scan.

I already restored the savestate(cal file) but still with error.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: re: re: GE HiSpeed FX/i problem

May 23, 2012 11:35

If service scan does not show error, then your hardware is ok...try loading whole SW, and do restorestate again.

Log inor Register

to rate and post a comment

Alexander Tecson

re: re: re: re: GE HiSpeed FX/i problem

May 24, 2012 11:47

the machine is now operational,. we found faulty cam board near at the reference position(channel).

Thank you for your support!

Log inor Register

to rate and post a comment

(1)

Arik Robart

re: re: re: re: re: GE HiSpeed FX/i problem

May 21, 2025 11:25

Hi, I am also facing same issue, but error channel is 291. Sometimes 290. Which converter card these channels are associated? Is FXI has NP DAS architecture? How to do Channel mapping here? I am facing vertical line while doing scout scan & finding ring in images.
Please support here...

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: GE HiSpeed FX/i problem

May 23, 2012 04:16

collimator problem

1 person liked this
Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE HiSpeed FX/i problem

May 24, 2012 11:47

thanks for the reply.

Log inor Register

to rate and post a comment