GE Hispeed RP (Hardware Stopped Scan Problem)

Zeeshan Sadiq

GE Hispeed RP (Hardware Stopped Scan Problem)

December 08, 2005 12:12

During routine scanning after first three slices a message appeared ("Scanner Hardware Stopped Scan. Wait 37 Seconds")and after which the machine didn't scan at all. What could have gone wrong?

Log inor Register

to rate and post a comment

Ralph Childs

Re: GE Hispeed RP (Hardware Stopped Scan Problem)

December 09, 2005 12:11

Zeeshan,

The HiSpeed RP delay is probably an initial tube cooling error due to the HV shutdown. Now, you need to diagnos ethe actual failure.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

(74)
(5)
(10)

Tayfun Canli

Re: GE Hispeed RP (Hardware Stopped Scan Problem)

December 09, 2005 02:05

Hi
System reports some problem about hardware, but I need details of message, please sent me error message details. Can you continue by turn off/on 550V and loop switches ( 3 switch at the right side of gantry cover) and hardware reset after that?? Is there any noise at gantry room or any error led lits on axil servo??

Log inor Register

to rate and post a comment

(18)

(14)

M.Naveed Ahmed

Re: GE Hispeed RP (Hardware Stopped Scan Problem)

December 09, 2005 05:39

Dear Zeeshan

Please cheque in support programme without X-Ray ON and let see the problem behaviour does change or not?

By the way where are you located?

Naveed Ahmed
Arrow-Tech Medical Systems Islamabad Pakistan.

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: GE Hispeed RP (Hardware Stopped Scan Problem)

December 09, 2005 10:15

You either have a high voltage problem or a data collection problem. You need to look at the error log and see what side of the system is being blamed for the error.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Certified Pre-Owned Endoscopes, All Major Brands, Best Pricing & Warranty.

We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and accessories. We are committed to your absolute satisfaction. (513) 469-7267

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)