Hello:
Recently reinstalled, after 6 months without use, a GE HiSpeed ZXi CT scanner had the following situation. After performing the Tube conditioning and calibration process, and after a few axial scans, everything without any problems or errors, we got the following errors:

XG Error, Error 186218
Phase is 5 code is 60
Detail Error Code 60-0317H
Exposure Error: Spit Ratio Error

Then every time we try to make a scan, either axial, stationary or scout and with any technique we got the following error:

XG Error, Error: 186218
Phase is 5 code is 60
Detail Error Code 60-0319H
Exposure Error: KV Unbalanced Detected

We performed the No Load Test, and without the High Voltage Cables connected to the HV Tank, we got a Heater Error and the 60-319H KV Unbalanced Detected Error, being this the error we're getting always.
I will appreciate any help that could be provided.
Thank you all.
Elias Valdes

Log inor Register

to rate and post a comment

Alan Ehrlich

Re: GE HiSpeed ZXi error

April 03, 2008 10:02

SEND US A EMAIL TO aje1214@yahoo.com

Alan Ehrlich

Log inor Register

to rate and post a comment

(16)

(2)

Ricardo Faria

Re: GE HiSpeed ZXi error

April 03, 2008 06:07

Dear Friend,
Please check error on Ct's HV tank.
The first error is usually any problem cause by sparks inside the x-ray housing. Normally, a X-ray tube revitalization will fix that.

Please check the schematics about the final stage before the x-ray tube, KV boards, ect.

Please let me know about your sucessful.

regards:

ENG. Ricardo Faria
EMG- Brazil

Log inor Register

to rate and post a comment

(57)
(3)
(6)

R.K.S singh

Re: GE HiSpeed ZXi error

April 04, 2008 01:32

dear
my self rohit singh CT ENGG I GOT SAME ERROR IN GE HISPEED FX/i I GOT SAME PEOBLEM I CHANGE TUBE AND JEDI BOTH HAVING PROBLREM Then i slove the problem if ur two parts r faulty then its deficault to diagnostic

Log inor Register

to rate and post a comment

(58)
(6)

Elias Valdes

Re: GE HiSpeed ZXi error

April 09, 2008 12:14

Thank you all, in fact it's difficult to diagnose if it's only the tube or the HV Tank or both.

Log inor Register

to rate and post a comment

(11)

toufik alkousa

Re: GE HiSpeed ZXi error

May 12, 2008 05:07

I have all manuals you may need for this system if you are interested plese contact me on: kousa@lycos.co.uk

Log inor Register

to rate and post a comment

(24)
(4)

Fabio Ferreira

re: GE HiSpeed ZXi error

February 24, 2014 10:05

i got the same error. solved you?

Log inor Register

to rate and post a comment

(66)
(9)

nilo panlilio,jr

re: GE HiSpeed ZXi error

February 27, 2014 01:12

replace xray tube and hv tank

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)