Over 150 Missouri Auctions End Today - Bid Now

GE Innova 2000 problem

Tags: Special Procedure, GE

(160)
(22)
(9)

Roy Nakamura

GE Innova 2000 problem

January 30, 2017 10:36

one of my customers are looking for solution of operational problem. he said it might have problem on Titan Assembly (P/N:2197212-Titan Assembly). how can it be fixed? Need to replace spare one? Thank you for your advise.

Log inor Register

to rate and post a comment

(54)
(9)

Mina halim

re: GE Innova 2000 problem

January 31, 2017 10:19

Hello Roy
I am specialized in that system.
What exactly the problem he faced??
What error code he had on the console??
I need to make sure that the problem is in the TITAN then we can fix it or replace it.
Waiting your reply.

Log inor Register

to rate and post a comment

(160)
(22)
(9)

Roy Nakamura

re: re: GE Innova 2000 problem

February 01, 2017 10:17

CPU of this Assembly is defective. Total itself is like independent PC. System can’t boot up without this assembly working properly.

Log inor Register

to rate and post a comment

(54)
(9)

Mina halim

re: re: re: GE Innova 2000 problem

February 02, 2017 10:02

So did you check the two LEDs in the TITAN??
What are their status??
Did the console touch scree power up or not (ofcourse with no data i mean the lighting of the scree).
After your reply i can know whether it is the CPU or harddisk or something else.

Log inor Register

to rate and post a comment

(1)

VarTom

re: re: re: re: GE Innova 2000 problem

June 19, 2017 05:06

Can someone help me with a Innova 2121 positioner error.Thanks

Log inor Register

to rate and post a comment

(48)
(9)

Francesco Didonato

re: GE Innova 2000 problem

January 31, 2017 10:20

Hello in tha TITAN there are two led One red and One yellow if during boot yellow led stay alwsys on You are problem on the titan you need to reload software normally led yellow off and green led on
Regards

Log inor Register

to rate and post a comment

(3)

Anthony Orozco

re: re: GE Innova 2000 problem

March 30, 2026 09:56

Hi Francesco, I have a question. Is it necessary to have a boot floppy disk in order to boot the Titan? If so, how can I create one? My TITAN ASC is 2197212, Innova 2000 GE

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: GE Innova 2000 problem

January 31, 2017 10:24

If you are sure that the Titan is defunct you have to replace it. But if there is a good computer Hardware engineer you can get it repaired. But get it confirmed from your customer that the Titan is bad.

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: re: GE Innova 2000 problem

February 02, 2017 10:06

Where is your location?

Log inor Register

to rate and post a comment

(160)
(22)
(9)

Roy Nakamura

re: re: re: GE Innova 2000 problem

February 03, 2017 10:10

in Europe

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: re: re: re: GE Innova 2000 problem

February 05, 2017 03:21

Hi
connect laptop serial port to COM1 of the TITAN which is in the top. Connect thro AdvantixE terminal. You should see the DOS prompt C:\>. Type DIR. you should see the files in HDD. if this test passes your CPU board, HDD etc are okay. The problem may be with BIU board or the Serial board inside the TITAN.
I am in India.

Log inor Register

to rate and post a comment