GE Innova 3100 detector elevator problem

Tags: Radiology - R/F, GE

(157)
(28)

Aleksander Proof

GE Innova 3100 detector elevator problem

January 11, 2015 12:33

Hello colleagues!
I have a problem with angiography GE innova 3100.
When I start to move the detector elevator from 120 cm down, the motion detector is interrupted at 118 and 116 cm. MCB board in the cabinet reboot after 5-10 seconds after that I re-push the button down and the detector continues to move down without any problems.
Problem occurs precisely when the detector will stand at the top of the maximum point of 120 cm for 2-3 minutes and when it begins to move down. When I move the detector constantly up and down, it does not stop

There are always same errors:1050120,1050045,1050049, 10500003.

I tried to change the MCB board together, i changed new potentiometer, i changed the used motor with an encoder, but that all does not help, and everything remains unchanged.
Please help!

Log inor Register

to rate and post a comment

(12)
(8)

Arturo Papa

re: GE Innova 3100 detector elevator problem

January 22, 2015 01:09

2.Make sure that the 24V supply voltage is present on the Brake input.
3.If voltage correct, check the Brake
4.If voltage is not OK, check the 24V power distribution
5.If the problem persists, check the FRT Lift axis MCB board

Log inor Register

to rate and post a comment

(32)
(2)

cesar lizarazo

re: GE Innova 3100 detector elevator problem

May 19, 2015 01:48

hello , can did you solve the problem.?

Log inor Register

to rate and post a comment

(157)
(28)

Aleksander Proof

re: re: GE Innova 3100 detector elevator problem

May 22, 2015 07:53

Yes
The problem was be in detector cable harness.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)