ge lemage sx/e image problem

February 08, 2009 09:41

Greetings! I have an image artifact with a ge lemage sx/e ct scanner. In relation to this I am having high values of sd data which averages at around 2.7, max of 5.4 and a min of about 0.9. I suspect a faulty adc board. I already replaced the capacitors on the adc hoping for bad filter capacitors but same problem, still high with the sd data. Thanks...

Log inor Register

to rate and post a comment

Charbel Tadros

Re: ge lemage sx/e image problem

February 26, 2009 04:58

try to do das transfer test.
if fail ,, look to the DBM Board .

Log inor Register

to rate and post a comment