My CT(GE Lightspeed 8 slice ,Octane2) get problem,
Only run warm up and shower "Stop Hardware stopped Scan", Check message list only Error:184154
Timeout on Exposure Command to X-ray on
HW Error:Possible Hardware problem
Timeout is:300 mS
If your system run warm up, your generator, HV Tank and Tube are OK. You must check in detail the error log. If you can download the error log, please send it to me: guery.diaz@maxialatam.com
steven
GE Lightspeed Error:184154
October 06, 2017 10:03
My CT(GE Lightspeed 8 slice ,Octane2) get problem,
Only run warm up and shower "Stop Hardware stopped Scan", Check message list only Error:184154
Timeout on Exposure Command to X-ray on
HW Error:Possible Hardware problem
Timeout is:300 mS
has any one can give some suggest
thank
steven
to rate and post a comment
VENGE
re: GE Lightspeed Error:184154
October 08, 2017 09:12
If your system run warm up, your generator, HV Tank and Tube are OK. You must check in detail the error log. If you can download the error log, please send it to me: guery.diaz@maxialatam.com
to rate and post a comment
Martin Martinez
re: GE Lightspeed Error:184154
October 08, 2017 09:14
Steven,
Please send me by email the error log pics.
to rate and post a comment