GE LightSpeed not scanning

Tags: CT, GE

(140)
(35)

FranciscoG

GE LightSpeed not scanning

September 18, 2015 07:36

Hi, I need some advice on this one, the system aborts scan everytime, and there are these errors in the error log:

- Error: 260118324
Host: On-board Rotating Processor
Function: XRAY GENERATION: Tube Rotor Control
Unable to bring up rotor due to acceleration interval constraint
Helical scan
Remaining time: 82 seconds

- Error: 260118192
Host: orp
Function: No system function reported
Jedi detected open filament circuit.

- Error: 260118363
Host: orp
Function: No system function reported
Rotor timers:
Rotor Off Delay Rem(mS): 0
Rotor On Delay Rem(mS): 0
Filament Off Delay Rem(mS):0
Filament On Delay Rem(mS): 0

Let me know what you guys think, I believe is the tube.

2 people liked this
Log inor Register

to rate and post a comment

Eric Hauk

re: GE LightSpeed not scanning

September 19, 2015 10:02

On Siemens scanners this problem can be caused by the generator control board. It's called a d400 in those scanners. I am just learning the light speed but I am sure they also have a similar control board. How many scan seconds are on the tube and what is the normal life on these light speed scanners

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: GE LightSpeed not scanning

September 24, 2015 05:47

Thanks Eric, I have a bad tube.

Log inor Register

to rate and post a comment

(43)
(21)

randy oswalt

re: GE LightSpeed not scanning

September 19, 2015 10:03

Which light speed system. 4.x 5x 7x

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: GE LightSpeed not scanning

September 21, 2015 01:19

Hi It is a 5x

Log inor Register

to rate and post a comment

(2)
(2)

Dzmitry Ramanchuk

re: GE LightSpeed not scanning

September 19, 2015 10:04

Disconect HV Cable from HV Tank and check resistance on the X-Ray Tube side small and large focus (between cmd and S & L connection). I think that you find open circuit - bad x-ray tube, replace it. If result is about 4 Oms the tube is OK and you have othere case.

2 people liked this
Log inor Register

to rate and post a comment

(14)
(7)

Syl Harris

re: GE LightSpeed not scanning

September 20, 2015 09:43

Tube has failed or rotor control (aux box) not likely though.

How old is tube or how many exams? Likely tube rotor bearings are bad. Was the rotor louder than normal before failure?

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: GE LightSpeed not scanning

September 21, 2015 02:45

Hi Syl, its the original tube 2007, the rotor wasn't noisy at all, just stopped scanning.

Log inor Register

to rate and post a comment

(5)
(1)

Rajesh Venewal

re: GE LightSpeed not scanning

September 21, 2015 08:33

If you want to solved this problem try to Change chopper circuit.

1 person liked this
Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: GE LightSpeed not scanning

September 24, 2015 12:03

Hi Francisco!. Sound like a tube failure. Follow the Dzmitry Ramanchuk troubleshooting. I'm sure the problem is the tube, even if the filaments are OK.

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: GE LightSpeed not scanning

September 24, 2015 05:47

Thanks, I do have a bad tube!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)