GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 01, 2012 10:06
Hi
My system GE LS.X3 has the following error
"the hardware reset failed". When I replaced the STC ARTESYN III board the error disappear and I got another error when I do KV&MA (X-RAY)TEST:
"Hard time: 86098 error type:firmware software error :error when posting to vrtx queue soft or hard reset may be required."
I appreciate any advise and suggestions.
Note: I do download flash without any errors.
I am waiting your reply.
to rate and post a comment
Mark Parris
re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 02, 2012 07:24
Did you change the jumpers on the Artesyn bd?
to rate and post a comment
Fadhl Alakwa
re: re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 04, 2012 08:44
Yes the jumpers are like the service manual configuration. You can not imagine that this error is because 2 broken resistors in the Axial board.
Thank you