GE Lightspeed table error 260134707

(157)
(28)

Aleksander Proof

GE Lightspeed table error 260134707

February 14, 2012 10:04

Hi!
On computed tomography GE Lightspeed VCT when I drop a table at the moment when I release the button or pedal a table immediately not stop and goes down in about 3 centimeters. In the error log:
error 260134707 Elevation could not stop with stop command. If table is moving downward, check valve.

Which valve should be changed: Proportional Valve or Shut Off Valve? Or it may have to do something else?

Log inor Register

to rate and post a comment

Eric Smith

Re: GE Lightspeed table error 260134707

February 15, 2012 05:46

The problem is that the Proportional Valve needs adjusting/replacing. Follow the procedure in the service manual to adjust the valve, careful not to turn the adjusting screw more than 20 degrees at a time. If you cannot get the table elevation down into spec - then I would replace the valve.

Eric Smith
SOS Medical Imaging
esmith@SOSMedicalImaging.com
Your CT Imaging Specialist!

1 person liked this
Log inor Register

to rate and post a comment

(66)
(9)

nilo panlilio,jr

Re: GE Lightspeed table error 260134707

March 24, 2012 11:44

verify or replace your elevation encoder. i had similar problem with this.

Log inor Register

to rate and post a comment

(157)
(28)

Aleksander Proof

Re: GE Lightspeed table error 260134707

March 26, 2012 01:02

I changed the Proportional Valve and table is ok! Thank you all for your help!

Log inor Register

to rate and post a comment