GE Logic 400

(2)

Elie MOARBES

GE Logic 400

March 11, 2011 09:28

I have the Logic 400 Ultrasound, when i turn off the machine it will restart again , cannot power it off unlees remove the electrical cord.
What shall i do ?

Log inor Register

to rate and post a comment

(46)
(9)
(1)

Victor Lupusor

Re: GE Logic 400

March 14, 2011 04:50

Verify BIOS,
reinstall software and application. I think scanner generate an error on shutdown procedure and perform a restart to fix it.
Regards,
Victor

Log inor Register

to rate and post a comment

(69)
(12)

marco antonio goncalez

Re: GE Logic 400

March 15, 2011 05:38

[13] [SHVST–PC] ERROR
1. Check the probes connected with console.
2. Check the DCON Assy or CONN1 Assy.
3. Check the Power Supply.

Log inor Register

to rate and post a comment

(14)
(5)

Israel Morales

re: GE Logic 400

January 27, 2015 10:10

The GE Logic 400 PRO Series, only turns on and off when you connect or disconnect the power cord, encende not now, when connected only listen as if to turn. Any help the grateful lot.