If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Ge Logiq 400 scsi error ( boot problem)
March 19, 2014 10:04
I have a ge Logiq 400 cl pro series when we switch on the machine machine show SCSI error.machine have 5.01 version
waiting any suggestion about this problem
to rate and post a comment
bens prado
re: Ge Logiq 400 scsi error ( boot problem)
March 19, 2014 08:58
CAn you give more details on the SCSI error? For the meantime try to check the SCSI cable connection, the power supply for your hard disk. Possible suspect is hard disk drive
to rate and post a comment
DigitalUltrasounds
re: re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 08:21
i checked all the things they are ok but when i change the hard drive with other machine.then machine work ok.i need support for solve this problem
to rate and post a comment
bens prado
re: re: re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 08:40
Your Hard disk drive(HDD) is bad. Replace your HDD and load the software. If you dont have the software copy the HDD of the other machine HDD to HDD copying.Bring it to the computer shop that is capable of duplicating your entire HDD.
to rate and post a comment
SHYJU MON
re: re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 08:24
Chek Low voltage power supply Distribution board which is having two fuses,lcated left side of LV powersupply remove fuses and make it connect directly, and see stil not solved your hard disk ,scassy card, scassy cable may be faulty
we have all these parts
shyjumon@hotmail.com
to rate and post a comment
Manuel Casanova
re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 10:38
send me you e mail for asm manuals
to rate and post a comment
only rocky
re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 03:01
hii is it scsi error 20166 max problum is scsi card which is below of the HDD, and also scsi cable chick once 5v and 12v in LV
to rate and post a comment
Zach Ruder
re: Ge Logiq 400 scsi error ( boot problem)
March 20, 2014 06:31
This does not always indicate a bad HDD. It can also be the CPU #12 Board which also has a scsi controller built into it. It can also only be a timekeeper issue as I have seen this before.
to rate and post a comment
DigitalUltrasounds
re: re: Ge Logiq 400 scsi error ( boot problem)
March 22, 2014 10:52
ok i will check
to rate and post a comment
MOTSO
re: Ge Logiq 400 scsi error ( boot problem)
April 02, 2014 07:44
I HAVE THE SOLUTION OF YOUR TROUBLESHOUTING mail me on theodmotso@gmail.com
to rate and post a comment
babusgajelli Gajelli
re: Ge Logiq 400 scsi error ( boot problem)
May 09, 2014 01:54
Pls check all kinds of bus cables&some ribben cables if you are all ready done this ignore it
to rate and post a comment