GE Logiq 9 OP Panel LCD Display Problem

September 30, 2013 08:22

I have a problem with GE Logiq 9 OP Panel (touchscreen) LCD display.
There is no image on the display I've reloaded the software, replaced the flat panel and the DVI cable but the problem still presists.
It seems that the PCI Graphics card or flat panel controller (LVDS to TTL converter) is damaged.
Any of them doesn't have GE P/N.
Could anyone supply that kind of parts?

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: GE Logiq 9 OP Panel LCD Display Problem

September 30, 2013 06:31

There are 2 types of toch screen depending on the BEP model, give the part number and I will let you know.
Contact me rey@allultrasoundservices.com

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: GE Logiq 9 OP Panel LCD Display Problem

September 30, 2013 06:31

what BTXX is this? if you touch the panel does it work?

Log inor Register

to rate and post a comment

(11)
(3)

Piotr Karczewski

re: re: GE Logiq 9 OP Panel LCD Display Problem

October 01, 2013 08:09

Every thing excep the display works fine.
System i s BT04 with BEP 3

Log inor Register

to rate and post a comment

(11)
(3)

Piotr Karczewski

re: re: GE Logiq 9 OP Panel LCD Display Problem

October 01, 2013 08:09

Everything except display works fine.
System is BT04 with BEP 3

Log inor Register

to rate and post a comment

(15)

Alexey Makeev

re: GE Logiq 9 OP Panel LCD Display Problem

October 04, 2013 09:24

what says system diaghostic?

Log inor Register

to rate and post a comment

(11)
(3)

Piotr Karczewski

re: re: GE Logiq 9 OP Panel LCD Display Problem

October 04, 2013 10:38

Nothing, no errors it says that the system is ok

Log inor Register

to rate and post a comment

(32)
(10)

Andrew Weydmann

re: re: re: GE Logiq 9 OP Panel LCD Display Problem

October 04, 2013 03:04

There is possibility that back light is off/burn out. Are you able to see anything using good flashlight (direct it near edge of LCD )?

1 person liked this
Log inor Register

to rate and post a comment

(11)
(3)

Piotr Karczewski

re: re: re: re: GE Logiq 9 OP Panel LCD Display Problem

October 07, 2013 08:56

I've replaced flat panel and inverter but nothin hapenned there is no power supply to the backlight from the BEP. I need PCI Graphics Card for further dgiagnisis.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)