hi,
I have a problem with imaging on all probes (are detected, options work). where is the problem: power supply, imaging board? anyone had such a fault?
Looking the images suggests it is a defective power supply, in its HV section.
Typical image of HV off.
Never seen Front Board failure before. Is it a possibility? I believe it is, but I'd bet in the HVPS.
Just in case: reinstall the SW all over. Let's imagine some part of main Ultrasound App, responsible to load boards' firmwares could be lost, provoking the HV to be untriggered?
We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!
Krzysztof Kuzniewski
ge logiq A5 - imaging problem
October 08, 2020 11:00
hi,
I have a problem with imaging on all probes (are detected, options work). where is the problem: power supply, imaging board? anyone had such a fault?
https://photos.app.goo.gl/hmfJt2jouXB8Rzg76
to rate and post a comment
Marcelo Antonio Correa
re: ge logiq A5 - imaging problem
October 09, 2020 10:14
Hi, Krzysztof.
Looking the images suggests it is a defective power supply, in its HV section.
Typical image of HV off.
Never seen Front Board failure before. Is it a possibility? I believe it is, but I'd bet in the HVPS.
Just in case: reinstall the SW all over. Let's imagine some part of main Ultrasound App, responsible to load boards' firmwares could be lost, provoking the HV to be untriggered?
I'd tried. It's easy and fast.
Hope these help, take care, God Bless.
Kind regards,
MAC.
to rate and post a comment
Paul Baker
re: ge logiq A5 - imaging problem
December 07, 2020 11:00
Hi there,
Did you find the source of this problem? I have a similar issue.
Thanks.
to rate and post a comment