GE LOGIQ P6 Issue
November 23, 2020 10:26
Hi, I am having an issue with a Logiq P6. Everytime I connect a transducer, I am getting an error code of "Abnormal TXHVH+ is detected. Please restart the system" Can anyone please tell me how to fix this?
to rate and post a comment
Joseph Asmaro
re: GE LOGIQ P6 Issue
November 24, 2020 10:03
Hello Rohit, restart your system, it might go away, if it doesn't then you most likely have an issue with the Power Supply Module (PST-009.)
to rate and post a comment
Ricardo A. Sanchez
re: GE LOGIQ P6 Issue
November 24, 2020 10:04
you may have 2 any of 2 problem's, power supply damaged or transducer is damaged and is making the power supply to fail