GE LOGIQ P6 Issue

November 23, 2020 10:26

Hi, I am having an issue with a Logiq P6. Everytime I connect a transducer, I am getting an error code of "Abnormal TXHVH+ is detected. Please restart the system" Can anyone please tell me how to fix this?

Log inor Register

to rate and post a comment

(42)
(6)
(7)

Joseph Asmaro

re: GE LOGIQ P6 Issue

November 24, 2020 10:03

Hello Rohit, restart your system, it might go away, if it doesn't then you most likely have an issue with the Power Supply Module (PST-009.)

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: GE LOGIQ P6 Issue

November 24, 2020 10:04

you may have 2 any of 2 problem's, power supply damaged or transducer is damaged and is making the power supply to fail

Log inor Register