Hi, If you don't have the save state MOD, you need to do the complete calibrations again. What is the error are you getting while loading save state MOD? Send me the image through whatsap at +91 94440 26607 or by email to shiju@tecsila.com
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
CRAIG WAMBURA
GE LS Plus kV cals
July 03, 2017 08:07
Dear all,
I have just re-installed software on a GE LS but when I try to do a restore state, the MOD can't restore the cal database.
I cannot perform any scans as the system says "No kV stations have been calibrated"
What can I do since I have no other Savestate?
to rate and post a comment
Pawel Kopysc
re: GE LS Plus kV cals
July 04, 2017 06:28
Make a calibration :)
Regards
Pawel
to rate and post a comment
Borhan Kalash
re: GE LS Plus kV cals
July 06, 2017 09:23
I do not have good experiance
but try to restore just the cal files.
whats up no. 00963954444992
to rate and post a comment
Shiju Karakkoottil
re: re: GE LS Plus kV cals
July 10, 2017 09:14
Hi, If you don't have the save state MOD, you need to do the complete calibrations again. What is the error are you getting while loading save state MOD? Send me the image through whatsap at +91 94440 26607 or by email to shiju@tecsila.com
to rate and post a comment