GE model Case Stress Test System Writer problem

(13)

(2)

Mike Brockhaus

GE model Case Stress Test System Writer problem

November 16, 2011 11:38

I have 2 each GE Case Stress Test Systems where the writer will spit out a blank sheet then will print the data; the left edge of the data is scrunched. The rest of the data and recording is fine. I'm thinking its the writer itself and GE no longer supports this model. Does anyone have an idea as to what is wrong?

Log inor Register

to rate and post a comment

Dean Stephens

Re: GE model Case Stress Test System Writer problem

November 16, 2011 11:56

Clean the print roller and the page sensor. If neither of these work, it could be the drive motor is on it's way out. With both acting the same way, I'm wondering if the paper isn't the correct type, also.

Log inor Register

to rate and post a comment

(13)

(2)

Mike Brockhaus

Re: GE model Case Stress Test System Writer problem

November 16, 2011 01:46

Thanks Dean,
I have gone into the writer utility and the test print is beautiful. One place uses Kendall paper, and the other place uses Marquette paper. I may give this a try as well (swapping paper)

Log inor Register

to rate and post a comment

Mike Filchak

Re: GE model Case Stress Test System Writer problem

November 16, 2011 12:14

Hi Mike,

I can help you with that. Give us a call at 1-800-866-6689 opt 1
We have been servicing GE/Marquette for 24 years. Try to call when your in front of the machine.

Mike Filchak

Log inor Register

to rate and post a comment

(13)

(2)

Mike Brockhaus

Re: GE model Case Stress Test System Writer problem

November 16, 2011 01:46

Thanks Mike,
The next time I'm in front of it (Probably next week) I'll give you a call. Thank you

Log inor Register

to rate and post a comment

(32)
(2)
(29)

Calixto Barja

re: Re: GE model Case Stress Test System Writer problem

May 29, 2012 11:59

Nice talk wit you today Mike, I cant send the system becouse I dont have to much time I need to send the equipment to Colombia in 3 days,please if you know who repair this in Florida let me know pleas
Regards
Calixto Barja

Log inor Register

to rate and post a comment

(5)
(2)

Kim Moore

Re: GE model Case Stress Test System Writer problem

November 16, 2011 06:38

Hi Mike, There should be a mylar strip on the metal bar inside the writer assembly that the capstan roller sits against when it's in the closed position. If the strip is missing, the CASE can intermittently skip pages & potentially track off center. If the strip exists, replacing it should help. Use part number 2015763-001.

Log inor Register

to rate and post a comment